Does anybody now how to manually calibrate the latest z-wave multisensor by AeoTech (Multisensor 6)? It has a manual calibration option for temperature, humidity, luminance and uv under configuration parameters, however it is not clear to me what kind of values have to be set here. Googling I found some suggestions, but none seem to work.

Thanx

Hi,

Did you figure this out? The parameters can be added in habmin. You would need a reading of the respective parameters that you trust to be able to add a calibration factor.

Hi,

Yes, I installed the latest zwave-binding.

Thanx,

Jesse

Hi,

for the manual calibration you have to edit the offset in parameter 201 - 204 in OpenHAB.

One Exable for the parameter 201 temperatur calibration. If the sensor read 18.5 °C and it is 20.1 °C you insert the value 16 in the parameter 201 ( a value of 10 = 1 °C, 100 = 10 °C). If the sensor reads a higher value you can correct it with “-” values too.

Luminance: if measured value is 800 lux and the standard value is 750 the calibration value is “-50”